Page 1 of 1

suitability setting in Empower

Posted: Wed May 25, 2011 5:07 pm
by vishnu
Hello guys,
I am using waters hplc with empower 2 for analysing metformin.

Acc. to monograph, the suitability check is, the resolution between melamine and metformin should be 10 by injecting Ref C (solution containing metformn and melamin).

How to get this in peak table on empower?
i tried in suitability tab in processing method but it is showing in 'e' form(Scientific) (for eg, 1.77e+001 for 17.7 resolution) i need it in numerical form. please tell how to fix it.

Re: suitability setting in Empower

Posted: Wed May 25, 2011 7:48 pm
by vofelyp
Hello Vishnu,

Left click on the header of the column in the peaik table. Choose Column Properties from the appearing list.
Unselect the exponential notification. You can set the name of the column, number of digit etc.
:D

Re: suitability setting in Empower

Posted: Fri May 27, 2011 1:37 am
by vishnu
Thank you for your reply