urgent! CO analysis Shimadzu 2014
Posted: Wed May 25, 2011 2:57 pm
by osmankaratum
Do you think can I measure Carbon Monoxide in Shimadzu 2014? FID detector and column is DB-624 (Agilent)
Classic column of GC.
Where can I find information about the method?
Thanks a lot.
Re: urgent! CO analysis Shimadzu 2014
Posted: Wed May 25, 2011 3:17 pm
by chromatographer1
You do not have the correct detector or column.
Usually an FID is not sensitive to CO. Using a DB-624 means that CO is not separated from many gases which could interfere.
You should be using a mole sieve column and either a TCD, or a Helium ionization detector, or a FID with hydrogen carrier and a methanizer column before the detector.
The application will depend upon the matrix you are injecting.
Do a little research from process control analyzer companies or hire a professional to help you.
Good luck,
Rod
Re: urgent! CO analysis Shimadzu 2014
Posted: Fri May 27, 2011 12:20 pm
by dpr
As Chromatographer 1 says, matrix dependant.
However we are doing ppm level CO in H2 using ppq pack column then onto a methaniser before FID detection.
What is the CO in?
Re: urgent! CO analysis Shimadzu 2014
Posted: Wed Jun 22, 2011 4:51 pm
by thohry
You can use FID with methanizer or TCD detector. I think you can use a Porapak Q Plot column for CO analysis.