Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
Hi,
I have to measure O2-N2 in NF3 with PDID detector using helium as carrier gas. What is the elution time of NF3 in a Molecular Sieve? Is the Molecular Sieve the best column to do this application?

Thanks for your support
I would not expect NF3 to elute from the column unless it is big enough to remain outside the pores in which it would elute at the void volume. Interesting analysis you have there.

Please let us know the result if you can, if 5A or 13X works best.

I would also try Carboxen 1000.

Can you do a backflush and just measure the O2 and N2?

BEST WISHES,

Rod
hi rod,
I'll go with a backflush and a 5A. After elution of O2 and N2, I'll flush out the NF3.

I'll get back to you with some results when I have it.

kind regards

dany
3 posts Page 1 of 1

Who is online

In total there are 29 users online :: 2 registered, 0 hidden and 27 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot], Semrush [Bot] and 27 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry