Luke Method in the Pesticides Analytical Manual (USFDA)
Posted: Wed May 25, 2011 2:01 pm
by lusi
Can someone how to do the calculations using the Luke Method in the Pesticide Analytical Manual of the USFDA?
We are being pressured to use this method but I have never done it before or have any calculations for it.
We have to use E1 or E5 (depending on sugar content) for extraction and also C1 for clean-up. But I have no calculations to apply to the reading from the MS or UV. So I am not sure if I get a reading of 2ppm from the instrument what that really means.
I would really appreciate some assistance.
Re: Luke Method in the Pesticides Analytical Manual (USFDA)
Posted: Tue Nov 22, 2011 3:39 pm
by mckrause
You're probably better off, depending upon the pesticide(s) you're interested in, using AOAC 2007.01 rather than the Luke MRMs. There are certain compounds where Quechers would not be the method of choice, but for most pesticides residues it is infinitely easier and faster than Luke.
Posted: Wed Dec 07, 2011 5:29 pm
by lusi
Thanks mckrause. Can you enlightening me what types of pesticides would not be good for QuEChERS. Also what exactly are polar pesticides?
Posted: Wed Dec 07, 2011 6:54 pm
by Don Shelly
Hi Luci!
Polar pesticides are pesticides that are soluble in water. Polar pesticides and quats like paraquat and diquat, ethephon, glyphosate, fosetyl-al won't extract using QuEChERS.
Best Regards,
Don
Posted: Wed Dec 07, 2011 6:56 pm
by lusi
Hi Don...Hope all is well with you...Thanks for the reply..I made a mistake I meant planar pesticides...I read some where about planar pesticides...I think...
Posted: Wed Dec 07, 2011 7:28 pm
by Don Shelly
I'm doing well Lusi! How are you?
Planar pesticides are compounds with ring structures, like DDT. They extract just fine, however they can be lost in the clean-up step if you use graphitized carbon black (GCB).
Don
Posted: Wed Dec 07, 2011 7:34 pm
by lusi
Dear Don. Thanks so much, I must have read it while looking at some QuEChERS kits. I am doing ok. Still waiting that training.
Thanks a whole lot for the info one the compounds that do not do too well with the QuEChERS method.