Page 1 of 2
transfer line from GC column to MS
Posted: Tue May 24, 2011 6:17 pm
by sepscientologist
Has anyone heard of using a meter or so of capillary between the end of the GC column and the inlet of the MS?
It is so annoying to have to break vacuum every time I want to put a different column on the GCMS. I'm
thinking that if there is a meter or so of capillary I could just remove the column from the capillary/gcms inlet
and quickly put a septum over it before putting on a new column.
Any thoughts?
Re: transfer line from GC column to MS
Posted: Tue May 24, 2011 7:09 pm
by Consumer Products Guy
Posted: Wed May 25, 2011 7:16 am
by Peter Apps
If you put a septum over the transfer capillary you will often get a little bit of rubber in the end, so you will have to cut off the tip before you can connect the column again. In the worse case the little bit of rubber will be small enough to get sucked down the transfer line into the MS
If you do want to cap the line, use a press-fit cap, I think that Restek do them.
Posted: Mon Sep 21, 2020 3:49 pm
by wf.callahan
I'm actually pondering the same question today. We frequently use a guard column as the "transfer column" from the capillary to the MS. It does save us time when switching between columns for different analyses. Does it matter what size ID the columns are? If I'm using a .25mm capillary column with a .32mm transfer column, will I encounter any chromatographic problems?
Posted: Mon Sep 21, 2020 8:59 pm
by James_Ball
Shouldn't be any problems, especially with larger transfer lines. I was using a metal Y connector to join two columns to a transfer line into the MS, but was having trouble getting the proper flows through the column. For that I had to use a 0.53mm transfer line with the two 0.18mm analytical columns, this allowed the end of each column to operate under vacuum instead of at an elevated pressure do to restrictions into the MS. With a single column it shouldn't matter.
The Restek adapter listed previously uses a 0.1mm transfer line to cause a restriction so that you can change a column without venting because it creates enough restriction that the flow into the MS is low when the column is removed, but I would still cool the source before using that method just in case.
I have also installed transfer lines before when I wanted the temperature of the interface to be higher than what the maximum temperature of the column phase was, to prevent the phase from bleeding into the source, that works good also.
Posted: Tue Sep 22, 2020 12:39 pm
by benhutcherson
As linked above, Restek and a few others sell "no vent" fittings for any MS that sort of rely on this principle along with a fitting that has a very tiny orifice through it combined with a really tiny capillary tube.
I was on the verge of ordering one, but then I read their tech literature on it. I wanted one for my Varian 300-TQ as at one time I was swapping columns frequently for different applications. I tend to prefer small ID columns anyway(.20 and smaller if I can get away with it) and I'd developed one particular method using a beastly 50mx.15m column. I think I was running a head pressure of around 90psi on H2 to get reasonable flow rates through it(and the...less than great...flow controllers on the 3800).
In any case, Restek's literature gives you a table to compensate for the no-vent device. IIRC, if you're running say a 20mx.18mm column, you end up telling the instrument it's something like 90m long. That's a LOT of pressure.
Posted: Tue Oct 13, 2020 2:58 pm
by BMU_VMW
We use this on al our GC-MSMS systems.
I put a pre-column (no phase, same diameter) in the transferline. This way there is no bleeding of phase due to the higher temperature in the transferline.
the connection between the end of the analytical column and the transferline is made with an ultimate Union and siltite ferrules.
https://www.agilent.com/en/product/gas- ... ion-for-gc
Posted: Tue Apr 20, 2021 1:42 pm
by fcozzi
Hello eveybody!
I would like to connect two 30mx0.25mm columns to the MS transfer line through an uncoated and deactivated restriction as well...
What diameter should have that restriction? And what about the length?
Posted: Tue Apr 20, 2021 5:57 pm
by James_Ball
Hello eveybody!
When I did two columns into one MS I had to use a 0.53mm line into the MS and then connect the 0.25mm columns to it with a Y connector. It works well and you don't have to do anything special with column lengths in the settings to get proper flows.
Posted: Mon Jul 19, 2021 2:21 pm
by fcozzi
Hello eveybody!
How long was the 0.53 mm uncoated column into the MS?
Thank you!
Posted: Tue Jul 20, 2021 4:21 pm
by James_Ball
Hello eveybody!
Thank you!
In my two column setup it was long enough that the splitter could sit on the floor of the oven, a gentle bend from the MS inlet and one at the bottom so the fitting was near the center of the oven.
Posted: Wed Jul 21, 2021 6:45 am
by fcozzi
Thank you!
Thanks!
Posted: Tue Jan 18, 2022 4:15 pm
by fcozzi
Hello eveybody!
Here I am again, which Y connector did you use, please?
I bought a Restek MXT but all three connectors have the same diameter:
https://www.restek.com/row/products/acc ... tors/3355/
In my case 0.25 mm as I use that diameter for my GC columns, It would be nice if it exists one Y-connector with the possibility to install i.e. two 0.25 mm GC columns and one 0.53 mm uncoated column to the MS...
My current setup works but the backpressure is kind of "delicate"
Posted: Wed Jan 19, 2022 8:00 am
by fcozzi
Hello eveybody!
I have found this:
https://www.trajanscimed.com/collection ... t-splitter
Does anybody have experience with this splitter?
Posted: Wed Feb 02, 2022 3:14 pm
by MSCHemist
I have a quickswap. It has a helium fed CFT plate with the column ends on the front with a ultimetal ferrule, and a 0.17m 0.1mm ID capillary on the other side going through the capillary. The helium pressure is kept at 4psi. It work well. It can be tricky to get a seal on the transferline threads.
The qualitative analysis GC/MS's have a helium fed 3 way splitter so that the flow goes to an FID, FPD or ODP, and MSD. We tried unhooking the column and some air did get in but it was negligible, it went from 2k nitrogen counts to 10k. I could probably turn up the helium pressure to eliminate that.