Method development

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
I am using Methanol as a diluent with 10000ppm concentration, (Compound is Acidic in nature) but peak shape is not proper even after increasing the buffer strength of mobile phase ...So Please let me know that How can I get Sharp peak?

regards,
Ajay
As seems to often happens, you have provided almost zero information that would be usefull in solving your issue. However, based on the limited amount of information provided it is likely one of 3 issues (or a combination of these).

1. Sample to concentrated (10000 ppm??)
2. Sample Diluent to strong (100% methanol)
3. Injection volume too large

Of course it could also be one of 10 (or more) other common problems that cause pook peak shape.

AA
What is Compound is Acidic in nature?
In nature many more acidic substancies - gastric HCL in stomach , citric acid in lemone, free thriphospate ATF in sperm... 8)
3 posts Page 1 of 1

Who is online

In total there are 21 users online :: 2 registered, 0 hidden and 19 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Baidu [Spider], John Guajardo and 19 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry