loic2810,
If looking at impurities in ethylene, you are probably looking at fairly light components (CO, CO2, acetaldehyde, acetylene, etc...) which is not where MS particularly shines over GC. If you are looking at trace impurities, TCD is not really the detector that shines for that kind of work since it usually sensitive down to 10's or 100's of ppm and I suspect you want to go lower than that.
Therefore, I would suggest GC with an FID keeping in mind Rod's comments about a methanizer if it will get you low enough detection limits for CO/CO2. If not, HID will certainly get you down low enough for CO/CO2 but the chromatography has to be clearly defined since the HID will also see everything else you might inject. This is not a simple system so getting professional advice (and probably a professional build) is good advice.
Best regards,
AICMM