I have HP 5890 series II which successfully works with Chemstation A 10.01. Now I tried to upgrade Chemstaion to B 04.01 but with this version can't esteblish stable connection with chromatograph. In Logbook Instr1.log every 3 second I see:

Code: Select all

2758 1da4 4da6a25f  300
5890         I/O error with GC: Switch to offline.          10:29:35 04/14/11
2759 1da4 4da6a25f  300
5890         Recovered from I/O Error: Back Online.         10:29:35 04/14/11
2758 1da4 4da6a262  300
5890         I/O error with GC: Switch to offline.          10:29:38 04/14/11
2759 1da4 4da6a263  300
5890         Recovered from I/O Error: Back Online.         10:29:39 04/14/11
and on hp 5890 display I see that parameters passes again and again... without stop.

new chemstation can't correctly works with hp 5890?

thanks,
-vadym
I have HP 5890 series II which successfully works with Chemstation A 10.01. Now I tried to upgrade Chemstaion to B 04.01 but with this version can't esteblish stable connection with chromatograph. In Logbook Instr1.log every 3 second I see:

Code: Select all

2758 1da4 4da6a25f  300
5890         I/O error with GC: Switch to offline.          10:29:35 04/14/11
2759 1da4 4da6a25f  300
5890         Recovered from I/O Error: Back Online.         10:29:35 04/14/11
2758 1da4 4da6a262  300
5890         I/O error with GC: Switch to offline.          10:29:38 04/14/11
2759 1da4 4da6a263  300
5890         Recovered from I/O Error: Back Online.         10:29:39 04/14/11
and on hp 5890 display I see that parameters passes again and again... without stop.

new chemstation can't correctly works with hp 5890?

thanks,
-vadym
Zdravstvi Vadim
1) http://www.anchem.ru/forum/read.asp?id=12288
2) Interlab in Moskow. Best to turn to the distinguished Boris P Lapin
Lapin@interlab.ru , He has friends in Kiev. Maybe will throw the program or manual.

То же но на русском.
Здравствуй.
1) На Анхеме терли эту тему. Один из вариантов- использовать Унихром.
2) Обратись в Московский интерлаб, interlab@ sevice.ru. Лучше всего к уважаемому Борис Петровичу Лапину, Lapin@interlab.ru. У него есть знакомые в Киеве, занимающиеся дистрибуцией и сервисом, подскажет телефончик. А может и поодкинет мануал или программный плагин из своих запасов.
If you look into the 'Install' folder of your ChemStation CD and read the 'readme.txt' file you will see that the 5890 is not supported in rev B.04.01.

Gasman
yes, I see that but there indicated:

Code: Select all


12.  5890A is no longer supported
---------------------------------

The 5890A is no longer supported by the ChemStation.
and I have hp 5890 Series II...

however, I find possible reason of problem - Min. Firmware Revision:

Code: Select all

5890 Series II	  |          A.03.02	
my HP5890 has only A.02.01 revision... as far I understand it's not simle task - upgrade firmware..?

thanks a lot,
-vadym
Updating firmware is do-able, but a little awkward. Details and firmware can be downloaded from the Agilent.com
unfortunately, I can't find firmware for hp5890 on agilent site and wherever in internet, only for latest chromatographs... seems agilent believes that hp5890 already hopelessly obsolete... :(

did you meet it elsewhere?

thanks,
-vadym
I'll be back at work Tuesday. If you post your E-mail address (e.g. abd -at-134.com) I can E-mail to you all the files I have already downloaded and saved. Some of these files I think are also on the ChemStation CD disc.
the reason of my problem was regional setting issue of my PC (with Ukrainian options). It should be set up to English (US), or, at least, following options:

• Decimal symbol = . (point)
• Digit grouping symbol = , (comma)
• List separator = , (comma)

seems as chemstation at startup unabled set some parameters it's called error which reconnect chromatograph again and again.

After choosing correct setting on PC HP5890 and Chemstation works like a charm!

Thanks a lot,
-vadym
I have HP 5890 series II which successfully works with Chemstation A 10.01. Now I tried to upgrade Chemstaion to B 04.01 but with this version can't esteblish stable connection with chromatograph. In Logbook Instr1.log every 3 second I see:

Code: Select all

2758 1da4 4da6a25f  300
5890         I/O error with GC: Switch to offline.          10:29:35 04/14/11
2759 1da4 4da6a25f  300
5890         Recovered from I/O Error: Back Online.         10:29:35 04/14/11
2758 1da4 4da6a262  300
5890         I/O error with GC: Switch to offline.          10:29:38 04/14/11
2759 1da4 4da6a263  300
5890         Recovered from I/O Error: Back Online.         10:29:39 04/14/11
and on hp 5890 display I see that parameters passes again and again... without stop.

new chemstation can't correctly works with hp 5890?

thanks,
-vadym
Go to www.5890.com
you may get some help, but I am not sure
:mrgreen:
9 posts Page 1 of 1

Who is online

In total there are 14 users online :: 1 registered, 0 hidden and 13 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 13 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry