Sudan Dye in Chilies
Posted: Wed Mar 16, 2005 5:37 pm
by Lynda L
Hello Everyone,
First of all, thank you all for your experience and advice! I am a frequent viewer of the forum, this is my first post.
Does anyone know of a HPLC method for the determination of Sudan I, II, III, and IV dye in chilie products? I have found a few methods for Sudan I.
I have both a Diode Array and a Fluorescence Detector.
I am also interested in column recomendations for this analysis. I currently have a YMC-ODS-AQ, Synergy Hydro RP, YMC-Pack Polyamine II, and a Zorbax-C8.
Any suggestions or reccomendations appreciated.
Lynda
Posted: Wed Mar 16, 2005 9:08 pm
by Uwe Neue
A standard reversed-phase column will be fine. I would start off with a gradient with phosphate around pH 2.5 and inject the 4 dye standards. You will find your peaks in a gradient from 5 % to 95 % acetonitrile. Once you know where the peaks elute, you can start the gradient with a higher organci composition etc. Next, you need to get the dyes out of the chili. You need to think about an extraction procedure.
Posted: Thu Mar 17, 2005 2:03 am
by Mark Tracy
I did Sudan I, II, III just for show one time. They are surprisingly hydrophobic. I used 50:50:0.5 MeOH:MeCN:AcOH on a 150mm C18 column, and they eluted between 3 and 6 minutes. (That eluent is useful for phenolic antioxidants, and I had some made up, so...)
As for the chili peppers, they have a nice assortment of pigments that are likely to get in the way. You will need your diode array.
Posted: Thu Mar 17, 2005 3:24 pm
by Lynda L
Uwe and Mark,
Thanks for the response. I will use both your info in my method development!
Lynda