I would like to completely remove the FID and use the outlet for a transfer line. I removed everything according to the service manual including the detector PCB (circuit board) but when I turned it back on the detector heating element was still getting hot. I had thought the GC would detect the removal and respond accordingly but it isn't. Is it enough to disconnect the block heater and sensor as well or are there other issues I should be aware of? The service and operating manuals I have do not describe a full detector uninstall or swap procedure.

Also, what would be the best software to use ultimately? The used unit came with a Dell PC (Win2000) with ChemStation Rev. A.10.02 but it constantly locks up. Would the latest version of ChemStation running on a newer PC be a good idea?

Many thanks for any advice.
NewChromatist,

Yes, you need to remove the heater and RTD from the connectors on the motherboard to finish the job. Otherwise the GC will still think the "detector" is present (at least from the heater perspective.) Removing these should finish the job.

Best regards,

AICMM
Is the screen saver turned off?

Version A10.02 will work with XP pro. I suggest you try that first. What HPIB card have you got in the PC?
3 posts Page 1 of 1

Who is online

In total there are 14 users online :: 1 registered, 0 hidden and 13 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 13 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry