why HPLC column oven doesn't heat
Posted: Tue May 17, 2011 1:24 pm
by AUlee
Hi all,
I came across HPLC column oven doesn't work recently. One day, HPLC was running on 85 degree C. When I came back after 3 hours and suddenly found that the temperature is only 60 C. It's keep decreasing and the oven heat function lost completely even though I restart it.
BTW, the column oven model "CTO-20A prominence" and is only used around 2 years.
what should be the problem?
thanks and have a great day!
lee
Re: why HPLC column oven doesn't heat
Posted: Tue May 17, 2011 9:43 pm
by JGK
http://www.ssi.shimadzu.com/products/pr ... ct=cto-20a
Given that the CTO20A has a maximum operational temperature of 85C, I would suspect that the heating control is not working.
Posted: Wed May 18, 2011 8:07 am
by LCbob
Not sure about this model but the earlier models used to run the cables through the door hinge, check to see if there is wear ( disclaimer not sure on this model ).
Posted: Thu May 19, 2011 12:53 am
by tom jupille
I've never worked with that particular unit, but that kind of failure has to come from one of three things:
- the heater is burned out
- the thermostat/controller electronics module has failed
- the temperature sensor has failed
or, a bad connection between these modules.
Per LCBob's suggestion, look in the "obvious" places for frayed cables. If nothing jumps out, open up the chassis and unplug/reconnect all the connectors (sometimes they vibrate loose). Check to see if there is a fuse (maybe in the heater circuit?) that has blown.
If there's no joy at that point, then it's a hardware failure. If you have another oven of the same model, try swapping the parts around (thermistor / heater / controller) to find out which is bad. If you don't then it may be cheaper to get a service tech in to look at it, because otherwise you're going to have to swap in new modules.
Posted: Sun May 22, 2011 2:47 pm
by AUlee
Thanks so much for your replies.
One technician from shimazu fixed it. He said it's caused by long time high-temperature using. One part of the device (probably a connecting part, I don't know exactly) become loose. When he push somewhere hard, the oven heats. But gently push, no heating at all. So, finally he tight somewhere. It works good till now.
Have a great day.
Lee