Sensitivity problem with halogenated pesticides

Discussions about GC and other "gas phase" separation techniques.

2 posts Page 1 of 1
I'm running two GC-MSMS systems from Waters (with Agilent GC and Gerstel PTV inj.). The oldest of them have had problems with sensitivity fore a long time now, and it is mainly the halogenated pesticides that is'nt sensitive enough. I've had engineers from Waters here that have changed the detector and one of the quads, cleaned the source and changed the column, but nothing seems to work. The other instrument are running perfectly with the same method and instrument parameters.
Now I'm thinking the problem is in the GC rather than in the MSMS. I wonder if there can be an active site in the PTV injector or somewhere else in the system that have this effect on halogenated compounds? Is there anything else I can try?
Hi
Calibrate system on sensivity. May be quickly- isotermic 250C , lindane only.
If big ( 1 mg/ml) concentration of pesticide good see- the MS good job , and need remonte (germetizated) GC- nut ferrules, ingection septas ...
2 posts Page 1 of 1

Who is online

In total there are 8 users online :: 1 registered, 0 hidden and 7 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 7 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry