NOx by GC?
Posted: Mon May 16, 2011 8:32 am
by Barbara
Hello,
Is it possible to analyse the concentration of NO in a sample by GC using FID detection? Or do I need to use TCD?
Barbara
Re: NOx by GC?
Posted: Mon May 16, 2011 11:06 am
by chromatographer1
Your best choice would be a Helium Ionization detector by AIC, or Valco, or Gow-Mac.
What is several thousands of dollars when you need the best, most accurate, and LOWEST numbers possible?
Want more?
A Mass Spec for tens of thousands of dollars?
Low temperature porous polymer packed column is the best column choice, but get the best packed, not the cheapest. It does make a difference. Ask for cleaned packing to be used when you order. Pay the few dollars more for much improved performance and fewer and smaller voids. Yes, into our lives a few voids must fall.
Have a good week,
Rod
Re: NOx by GC?
Posted: Mon May 16, 2011 2:50 pm
by AICMM
Barbara,
Depends a lot on the level. If hundreds of ppm then a TCD would work fine I think. If low level or really low level then I would recommend a PID or an AID (argon ionization detector.) (The standard disclaimer applies here, I make both of these detectors.) With an IP of about 9.2 both of these detectors would have good sensitivity for low level NO. I am not sure the impact of the window in the PID but this compound has been done by AID and I can post a chromatogram if you would like. You will need to watch out for CO2 as it may be a potential interference. The other thing to be aware of is reactivity and this is one of the big reasons to pay attention to column quality.
In this particular case, I would not recommend an HID since you will have to have very good separations since the HID will see everything else you throw at it. I often state that an HID is like a TCD on steroids, it will see everything but with much stronger signals (lower detection limits.)
Best regards,
AICMM
Re: NOx by GC?
Posted: Tue May 17, 2011 7:54 am
by Barbara
Thanks so much for both your replies! Much appreciated.
Barbara