Alternative to Glycine buffer (pH 9.0)
Posted: Sun May 15, 2011 12:19 pm
by solntzevorot
Hi all,
I'm searching for glycine buffer alternative for mobile phase.
I'm using gradient method : sol.A Glycine buffer (pH 9.0), sol.B : Acetonityl:MeOH 85:15.
I'm searching appropriate buffer at pH 9.0 for UPLC system (but not Ammonium based).
Please help
Re: Alternative to Glycine buffer (pH 9.0)
Posted: Sun May 15, 2011 4:48 pm
by Kostas Petritis
Does it need to be volatile? If not, how about Tris?
Posted: Mon May 16, 2011 5:22 am
by solntzevorot
No, it doesn't have to be volatile.
What about phosphate buffer, prepared from tri-basic and di basic phosphate?
Posted: Mon May 16, 2011 5:36 am
by Kostas Petritis
It should work as the pka's are 7.2 and 12.32, barely within the +/- 2 "rule".
Posted: Mon May 16, 2011 6:54 am
by solntzevorot
Posted: Mon May 16, 2011 6:58 am
by Gerhard Kratz
Another organic buffer is CAPSO, pKa 9,7.
Or 1-Methyl-piperidine
Or Borate, but it is not volatile.
Posted: Mon May 16, 2011 8:15 am
by DSP007
Hellow
Volature basic buffer
1) ammonia buffer (NH3/+H20/+ NH4CL) or NH3/+H20/+ (NH3)2CO3
This russian langvige , but text scool task understood without translation
http://forum.xumuk.ru/index.php?showtopic=72307
2) triethylamine buffer (C2H5)3N+H20 + CH3COOH
Borate buffer -good stabile buffer , but in LC-MS borate buffer must not use.
Posted: Mon May 16, 2011 6:27 pm
by unmgvar
can you say why not ammonium based buffer?
Posted: Tue May 17, 2011 8:29 am
by DSP007
Nasty (Ungly) it is.
Volatile, malodorous, unstable and RP-silica gel "eats". But because of volatile buffers for UPLS-MS the ammonium is most volatile.
2 solntzevorot (Midsummer)=> by nick can be seen from the "brozer slavjans". Come on (part of) on anchem.ru
Posted: Tue May 17, 2011 8:32 am
by HW Mueller
Am I going bananas or did solntzevorot say his buffer need not be volatile?
Posted: Tue May 17, 2011 9:37 am
by DSP007
Hello HW Mueller.
Very likely when you worked with the ammonia buffer and know that
"it's not sugar".
Posted: Thu May 19, 2011 4:58 am
by solntzevorot
Thanks to all for the replies.
The buffer no need to be volatile.
Whith ammonium buffers (at least with those I have in the lab) I experienced with to much system peaks in the UPLC system.
to DSP007 : thanks for the link, I freely speak Russian and Ukrainian. (Born in USSR
Posted: Thu May 19, 2011 5:39 am
by Kostas Petritis
You shouldn't experience system peaks with ammonium buffers. If you saw peaks is probably either the low purity or contaminated salt. I would try again with some new high purity container.