Need help understanding retention time differences....
Posted: Tue Mar 15, 2005 7:59 pm
I'm attempting to run a residual solvent method that was developed for GC/FID on an HP6890 w/ HP5973 MSD.
The method in question has been run on a 6890 with FID several times before. The analytes are methanol and chloroform. On the FID instrument, methanol typically elutes @ ~2.2 minutes and chloroform @ ~4.2 min.
When I run the same method on the MSD, methanol is eluting @ 1.1 min and chloroform @ 2.2.
I've confirmed that the oven temperature program is the same, and the oven temp is accurate throughout the run. I've also tried more than one column. Retention times remained the same.
The instrument parameters are as follows -
Oven program:
T °C / Hold, min / Ramp, °C/min
50 / 3 / 20
150 / 0 / 30
300 / 10 / NA
Flow: Constant, 2.0 mL/min
Injector:
Temp = 200 °C
Carrier: Helium, split ratio = 20:1
Injection volume: 2 uL
Aux temp: 280 °C (Original method has FID temp set at 300 °C)
Column is a Restek Rtx-5, 30 m, 0.32 mm i.d., 1um film
Run Time: 23 minutes
The sample solvent is DMF. I know this isn't the best solvent to use for MS applications, but my hands are tied at this point. That said, the DMF peak elutes in just under 5 minutes on the MSD but is not typically observed when using FID because the method states that data should only be collected for 6 minutes. . . Another clue that everything is eluting too quickly on the MSD.
My question is this: Has anyone observed such a drastic decrease in retention when switching from FID to MS before? Is this behavior inherent to the technique (due to vacuum pulling at the detector end), or is some other explanation more feasible?
Thanks for any feedback.
-Joe
The method in question has been run on a 6890 with FID several times before. The analytes are methanol and chloroform. On the FID instrument, methanol typically elutes @ ~2.2 minutes and chloroform @ ~4.2 min.
When I run the same method on the MSD, methanol is eluting @ 1.1 min and chloroform @ 2.2.
I've confirmed that the oven temperature program is the same, and the oven temp is accurate throughout the run. I've also tried more than one column. Retention times remained the same.
The instrument parameters are as follows -
Oven program:
T °C / Hold, min / Ramp, °C/min
50 / 3 / 20
150 / 0 / 30
300 / 10 / NA
Flow: Constant, 2.0 mL/min
Injector:
Temp = 200 °C
Carrier: Helium, split ratio = 20:1
Injection volume: 2 uL
Aux temp: 280 °C (Original method has FID temp set at 300 °C)
Column is a Restek Rtx-5, 30 m, 0.32 mm i.d., 1um film
Run Time: 23 minutes
The sample solvent is DMF. I know this isn't the best solvent to use for MS applications, but my hands are tied at this point. That said, the DMF peak elutes in just under 5 minutes on the MSD but is not typically observed when using FID because the method states that data should only be collected for 6 minutes. . . Another clue that everything is eluting too quickly on the MSD.
My question is this: Has anyone observed such a drastic decrease in retention when switching from FID to MS before? Is this behavior inherent to the technique (due to vacuum pulling at the detector end), or is some other explanation more feasible?
Thanks for any feedback.
-Joe