Loss of Upper End of Linear Range - AB 3200
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
1 post
Page 1 of 1
I am running a method for EDDP with a normal linear range of 5ng/mL up to 1000 ng/mL on-column. Chromatography is an Acquity using a Phenomenex Kinetex column and the usual formic acid in Di water/methanol mobile phase. We use a single-point calibrator at 30 ng/mL and verify the linear range every 6 months. Right now I am having difficulty with the top point having -25% to -40% deviation. This seemed to point to a saturation effect, but decreasing the injection volume, diluting the samples, and even detuning the MS settings for that compound have had no effect on this other than to cause the low points to become nasty. :/ Any suggestions would be quite welcome! Thanks!
1 post
Page 1 of 1
Who is online
In total there are 46 users online :: 0 registered, 0 hidden and 46 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 46 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 46 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science