separation of carbohydrates by HPLC using RID detector
Posted: Tue May 10, 2011 2:31 pm
please i need a help
hi, i am new to HPLC and i am working on the detection of sugars in honey mainly fructose , glucose and sucrose,
i am using a specialized column for carbohydrates from Agilent, with RID detector.
the HPLC is new and automated one from Agilent,
so this is the first trial of the column and detector.
i used to prepare standards as the following:
2 g fructose with 25 ml methanol up to 100 ml water
1.5 g glucose with 25 ml methanol up to 100 ml water
0.250 g sucrose with 25 ml methanol up to 100 ml water
but the peak of fructose used to continue under the base line before returning back, so i prepared the same standards without methanol and it seems that the problem is fixed but not knowing why?.
i prepared 2 honey samples as 5 g honey up to 100 ml water.
these numbers are as the method of bogdanov.
i used 80:20 ACN:water as mobile phase, with flow rate 0.9 ml/min.
the result i got is as the following:
1_when using the fructose standard i got a good peak at about 3.4 min
2_the glucose standard showed exactly the same peak as the fructose!!!!!!!!!! and another small peak at about 6.9 min ,
3_sucrose standard showed a perfect peak at about 9.1 min , but still the fructose peak appearing..
Is this a fructose peak????????
and why it is appearing in the 3 standards, and if not where is the fructose peak????????
the surprise is that the standard mixture showed fructose peak , glucose peak but sucrose peak did not appear at all?????????????
this vial i prepared by taking 0.5 ml from each standard by micro-pipit.
i run the honey samples in the same run sequence but they did not show glucose at all and this seems not logic some how , but they show fructose and its calculation was very logic(about 32 g\100g).
i am mixed from these results,
please help
what shall i try or do to get a good results.
hi, i am new to HPLC and i am working on the detection of sugars in honey mainly fructose , glucose and sucrose,
i am using a specialized column for carbohydrates from Agilent, with RID detector.
the HPLC is new and automated one from Agilent,
so this is the first trial of the column and detector.
i used to prepare standards as the following:
2 g fructose with 25 ml methanol up to 100 ml water
1.5 g glucose with 25 ml methanol up to 100 ml water
0.250 g sucrose with 25 ml methanol up to 100 ml water
but the peak of fructose used to continue under the base line before returning back, so i prepared the same standards without methanol and it seems that the problem is fixed but not knowing why?.
i prepared 2 honey samples as 5 g honey up to 100 ml water.
these numbers are as the method of bogdanov.
i used 80:20 ACN:water as mobile phase, with flow rate 0.9 ml/min.
the result i got is as the following:
1_when using the fructose standard i got a good peak at about 3.4 min
2_the glucose standard showed exactly the same peak as the fructose!!!!!!!!!! and another small peak at about 6.9 min ,
3_sucrose standard showed a perfect peak at about 9.1 min , but still the fructose peak appearing..
Is this a fructose peak????????
and why it is appearing in the 3 standards, and if not where is the fructose peak????????
the surprise is that the standard mixture showed fructose peak , glucose peak but sucrose peak did not appear at all?????????????
this vial i prepared by taking 0.5 ml from each standard by micro-pipit.
i run the honey samples in the same run sequence but they did not show glucose at all and this seems not logic some how , but they show fructose and its calculation was very logic(about 32 g\100g).
i am mixed from these results,
please help
what shall i try or do to get a good results.