I´m having some problems to analyze Ethyl acetoacetate by CG.The supplier of this solvent has sent me their analize method but it doesn´t work when I have tried.I am obtainig a peak with a big tail and bad resolution.The method of my supplier is the next:
HP-1 30m, 0,25mm,0.25µm T injector=180ºC Tdetector=250ºC
Flow=2 ml/min(hydrogen) split flow=200ml/min
T initial=50ºC(1.5 min) Rate=40ºC/min Tfinal=200ºC(4.0 min)
I only have to my disposal the next coloum:
HP-1 30m, 0,32mm,5µm and I have to work with Helium instead of Hidrogen due to the gas istalation only let work me with this gas.
Could be the poor resolution caused by these two factors?
Thanks