peak area percent change by isocratic elution
Posted: Fri May 06, 2011 1:52 pm
Hi, everyone!
Recently I am developing a isocratic elution mehod for a diastereomer. There are two peaks could be separated on ZORBAX SB-CN column, but peak shape isn't very well. The strange thing is peak area percent changed with the ratio of mobile phase.
One is 0.2% H3PO4 in H2O : ACN = 70:30. Resolution between two peaks is about 1.6 and tailing factor is about 1.8. Peak area percent of two peaks is about 70%:30%.
The other is 0.2% H3PO4 in H2O : ACN = 72:28.Resolution between two peaks is about 1.8. Peak area percent of two peaks is about 80%:20%.
Could you tell me what is the reason?