quite useful. When I compare integrated peak areas at 280 nm
with peak areas at 214 nm for different proteins using calculated extinction coefficients
according to this paper or Pace et.al (1995), resp. , I get pretty accurate and similar results, w/o
using a calibration curve. (The equation from Buck et.al 1989 is rather useless.) So if you
are dealing with a lot of different proteins and would like to estimate puritiy and quantity
after purification by HPLC, you may find this paper useful as well. The interesting thing is,
that the extinction coefficients at 214 differ quite remarkably between proteins (>30%!). So
just using BSA for calibration may not be very helpful. And I found that the determination of
the so called instrument response factor is indeed unnecessary if you use a modern
detection system reporting (m)AU instead of (m)V and calculated coefficients, as one
would assume. (You don't calculate response factors for you spectrophotometer, do you?)
But since I did this only with a small set of proteins so far I
would be interested in the experience of other users in this forum.