HP 6890/5972A GC-MS: "cannot bring up GC" error

Discussions about GC and other "gas phase" separation techniques.

5 posts Page 1 of 1
Hi everybody,

I am using an HP 6890/5972A GC/MS w/ Chemstation GC1701BA version B.01.00. The software was recently reinstalled and now will not connect to the instrument. The interface is an Agilent 82350B PCI GPIB. Thanks in advance for any help you can provide!

-Pat
Did you configure the GPIB board as "hp82341" even though it is an 82350B ? You need to. And change the bus address to 30. You need to.
Thanks very much, that fixed it. The problem was an incorrect SICL name and bus address. You guys are a great help!
Thanks very much, that fixed it. The problem was an incorrect SICL name and bus address. You guys are a great help!
REALLY????

Well, great - Because I have no idea what a SICL name is, and to me a bus address is where your old VW Type 2 is parked. And isn't a router used in the woodshop?
Isn't software a type of comfortable shoe?
GCguy
5 posts Page 1 of 1

Who is online

In total there are 27 users online :: 1 registered, 0 hidden and 26 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Baidu [Spider] and 26 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry