I am collecting 4 channels (2 UV/VIS wavelengths and 2 fluorescence emmissions = 4). I am able to collect all four channels with each injection (just trying to calibrate now). But that is where I am stumped. I have no experience with Empower. Do I need different processing methods for each channel or can I do it with one?
-
- Posts: 25
- Joined: Mon Aug 30, 2004 2:08 pm
You can process multiple channels with only 1 processing method, as long as all the events and RT windows work for all of your detectors. Typically, with 4 detectors in line, the Rt of your components will be increasingly longer. As long as the resolution of your components is sufficient, you can make your RT windows big enough to be able to correctly identify the peaks. Another consideration is the noise for a fluorescence detector usually going to be greater tha a UV detector, so you may want to ensure that the integration is good for all the chromatograms. All you need to do to process all 4 channels using the same proc. method is to put all 4 channes in your method set and indicate the appropriate processing method for all.
-
- Posts: 74
- Joined: Sat Jul 18, 2009 4:33 pm
As suggested, you can use a single processing method if this works or you could also use a different processing method for each data channel...,
To account for the different retention times from channels from different detectors (different tubing lengths!) you can use the filter/ offset tab of the derived channels in the method set or in each processing method (if different processing methods are used) and set an RT offset to normalise the channels.
Hope this helps,
Ben
To account for the different retention times from channels from different detectors (different tubing lengths!) you can use the filter/ offset tab of the derived channels in the method set or in each processing method (if different processing methods are used) and set an RT offset to normalise the channels.
Hope this helps,
Ben
3 posts
Page 1 of 1
Who is online
In total there are 16 users online :: 0 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 16 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 16 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science