ATP by LCMSMS

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

2 posts Page 1 of 1
I had following gradient and column for ATP detection.

Colum: waterxterra,C8,2.1x50mm.
5micron

Mobile Phase A:10mM DMHA( made by mixing 1.73 ml of DMHA(mw:129) in 998ml water) Adusted the Ph to 7.0 by dil HCl.
Mobile phase B: Methanol

ATP was made in water 50uM

I see ion at 130(ESI+VE) and 295(ESI+VE) all over the run more than 10 million count. nothing else. UV-VIS shows ATP comes out at 1.5 min. but dont see any in ion at 508 or neg 506 in -ve mode.

What happed to mass spec ? is 130 is due to DMHA..i dont understand what is 295 ?

Please help
Hello,

sorry for seeing this so late.

You might have degradation of phosphate moeities on heated steel surface and subsequent leakage. I have not checked how this is linked to the 295 m/z signal.

Suppression effects of carbonate on the interaction between stainless steel and phosphate groups of phosphate compounds in high-performance liquid chromatography and electrospray ionization mass spectrometry
Asakawa Y et al
Journal of Chromatography A
Volumes 1198-1199, 11 July 2008, Pages 80-86
2 posts Page 1 of 1

Who is online

In total there are 23 users online :: 0 registered, 0 hidden and 23 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 23 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry