LOQ limit impurities
Posted: Thu Apr 28, 2011 1:17 pm
In the european pharmacopee, one of the system suitability items is the control of the s/n ratio of the disregard limit by the determination of related substances. The s/n of the disregard limit must be >10:1!
Now I work at an generic farmaceutical laboratory and in the specified document for an tablet product, there is only the LOQ described. The determination of related substances is based on the total area of all impurities + the area of the principle peak. The known impurity is calculated as follows:
(area imp * 100%/((sum area imp)+area API))*Factor
The pre-determinated LOQ is <0.02%
When I calculated a sample, the known impurities has an percentage of 0.04%
This peak has an s/n ratio of 4. This is to low in observing the SS-limit of minimum 10 LOQ.
The method is set up by an other laboratorium, with possible an sensitiver detector.
Is it correct to control the s/n ratio here? Or because of no information in the analyse method, I just close my eys
. Anyway I can not do this
Can I, within the requiremends of the european pharmacopee and ICH guidlines, raise the injectionvolume so that the s/n ratio will be >10:1 ?
I can only find something about decrease the injection volume.
When I rais the injectionvolume do I have to validate this method?
Sorry for the many quenstions
, but I can not conform this with some one else within my company
Thanks a lot
Ing
Thanks a lot
Ing