I am currently developing a gradient method for a related compounds analysis. I am using an Agilent 1100 equipped with a DAD at 215 nm. My method utilizes a Waters XBridge C18 150 x 4.6 mm column with 3.5 um packing. The gradient is not linear but basically progresses as follows:
0 minutes - 70% K2HPO4 (pH 8.0) / 30% Organic (I am using both ACN and MeOH to achieve necessary selectivity)
28.5 minutes - 40% K2HPO4 Buffer (pH 8.0) / 60% Organic
I follow this with a 3 minute hold at 70% organic to rinse any "junk" off of the column and then reequilibrate at initial conditions for six minutes.
I have observed a gradient peak in blank and sample injections that elutes at ~ 22 minutes (~54% Organic) and interferes with a degradent of interest. After thoroughly washing the system and column, the peak is not present in early injections but consistently grows throughout a sequence and is much larger than my method LOQ by the last injection. In subsequent sequences in which no system wash was initially performed, the peak is present in early injections and continues to grow until the area stabilizes after several injections. The peak appears even if I run the gradient with no injection. I have tried multiple lots of water, K2HPO4, MeOH, and ACN. I have also switched instruments and columns. The peak is still present. The only thing that did help (somewhat) was switching from an HPLC grade ACN to a UV grade.
Based on these observations I do not feel the gradient peak is produced by my buffer. If it were, I would expect the peak to be large in my first few injections and then decrease in size as it is rinsed off of the system/column by my gradient. I am confident that the source of the peak is the ACN...this makes sense to me since I notice the peak increase in size as I continue to rinse with a high ACN content at the end of each injection (essentially building up more of the contaminent on the column). Am I misguided here?
If the ACN is the source are there any steps I can take to clean up the ACN before running my method? Would eliminating ACN from the column rinse at the end of each injection and using MeOH instead help?
Thanks...any input is greatly appreciated