Best guess in the absence of any detail at all, is that the peaks are tailing as they come out of the inlet, and that they are sharpened up by thermal focussing on the relatively cooler column. If the column is hotter (which is what I presume you mean by initial temperature) then the thermal focussing effect is weaker, and you still see the tail on the peak at the end of the column.
Thanks a lot Peter. I think the answer is just what the prof told me told he didn't use the phrase "thermal focussing".
Sorry about not giving the detail. I had too much going on that day with a car accident two days earlier in the evning - I am not making an insurance claim because I hit a curb just at the right angle and hence really hard). Picking up the car today.
It was the first day of this GC analysis of derivatized fatty acids (C12, C14, C16, & C18), namely methyl esters. In the 1st analysis, I pretty much the temperature program the way the prof had shown us as demo, which was initial temperature of 100ºC (hold for 2.5 min), ramping 30ºC/min, final temperature 250 (hold for 3 mins). Total run time 10.5 min. Injection was splitless and inlet heater temp was set at 250ºC; Septum purge flow was 3ml/min. If I remember correctly, professor said that the injection was no direct-on-column approach. We were told to leave the pressure setting alone. I will record it today. last time, I was busy familiarizing myself with the instrument and then rushed to return a rental car in time.
To reduce some time, I raised the initial temperature to 140ºC and peak tailing occurred. Peak tailing for very large for the 1st peak, pretty big for the second and small for the 3rd and 4th peaks.
Some feedback to the forum on how you are progressing with the other projects that you sought advice on would be nice.
Oh, yes, Peter. All lab work (1. Intro to HPLC, 2. Intro to GC, 3. SPE extraction+ HPLC analysis) had gone well. Currently working on Derivatization + GC analysis (lab 4). The professor is very strict abut the report writing though. Lab 3 is due next Monday.
As for the specifics w/ HPLC instrument (during lab 3), on last day which was last Tuesday, the ret time shift happened again, obviously due to eluent composition running through the column. Due to time limit for use of the instrument, I just let it be since the one peak I was analysis at different conc: came out during the analysis.
For the current lab, I do have another question that I didn't get to ask the prof that day whether or not the inlet heater temperature should be the same as initial temperature of the temperature program. I will ask him today.