by
orsfph » Tue Apr 26, 2011 5:34 pm
Hi, Peter,
Yes, the samples will dissolved in phosporic acid. You means the acid will cause damage to syringe etc? But for FID test, is it required to acidify samples to a certain pH, say pH 2, before injection?
I got this method from a journal paper, Bioresource Technology, 2009, 100, 4133-4138.
Anyway, my plan is to test the VFAs of effluent from a digestor using FID with a capillary column. Could you please refer me a detail procedure?
Thank you very much for your kind help!
If I understand correctly your samples will end up dissolved in phosphoric acid ?
This will destroy the syringe, inlet and column. Do not do it. Where did you get this method from ?
What column are you proposing to use ?, with a robust packed column you could make aqueous injections (but not with a non-volatile inorganic acid dissolved in them), but on a capillary these will cause all sorts of problems.
The "carbon number" of acetic acid and acertone are irrelavent, a wax type phase will easilt separate acetic acid from acetone.
Peter