Noise in PDA at 190 nm to 230 nm

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

6 posts Page 1 of 1
hi,

We have a Shimadzu VP series HPLC. Recently, the PDA has been creating a lot of noise in the 190nm to 230 nm. The lamp is brand new.
 

Please find attached the output we generated running 60%ACN/water at 50uL/min.  We also looked at the noise in the flat parts of the chromatogram at different wavelengths and would be interested in your comments.  There is a column in place but it has been thoroughly washed after the last few days so don’t think it can be to blame.

 

Wavelength       noise

 

190nm              3mAU

200                   3.5

210                   3

220                   0.7

230                   0.1

240                   0.02

[/img]
angela

Mark "excellent" :wink:
I have 'same' problem: wavy baseline at 200 - 220 nm. The mobile phase contains ACN - MeOH - water (bidestillated-demineralized water).
I do not know why ... :roll:

It is due to air bubbles in your pump. Open its vavle and purge the eluent throgh the pump.

Yeah... I hope the problem 'just' bubbles not in the electricity. :)

A "wavy" baseline with an identifiable period may be a pump problem. Is the period the same as your pump cycle? If you are using a high-pressure mixing type pump, the period will change with the %A/%B. Lots of things, including a bubble, dirty solvent, bad checkvalve, or contaminated pump head can do this.

A "noisy" baseline that is just random probably means that there is not enough light getting through. Is this a new problem, or has the detector always done this? It might be a dirty flow cell. Some micro-cells just don't pass that much light, and you can't fix that. Excess absorbance in the mobile phase can also do it, as can column bleed.

In a truly bizzare case, I once saw corroded gratings. The poor operator had two detectors with this same problem, and was really confused when a detector swap failed to fix the problem.
Mark Tracy
Senior Chemist
Dionex Corp.
6 posts Page 1 of 1

Who is online

In total there are 14 users online :: 1 registered, 0 hidden and 13 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Semrush [Bot] and 13 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry