Agilent 1200 SL binary Pump - Rusting
Posted: Mon Apr 18, 2011 5:31 pm
Hi All,
Does anyone have experience with Agilent pumps rusting? I saw one post here in 2008. We've had two pumps now with issues.
I am seeing a reddish discoloration on our pump head, outlet check valve threads. We also are getting a lot of leaks from the pump and AS capillaries. I do not recall ever having to change a stainless capillary as often as I have been this past year (at least 30 times - same and different caps).
We are running fairly conc. TFA (per Agilent Method Dev. folks) - 0.1% TFA/H2O (pump A)and 0.4%TFA/AcN (pump B).
Agilent has replaced one pump with a demo unit and after 3 months this one is experiencing leaks (3 in the past month). As we are under service contract.
Agilent has been extremely slow to respond - and do not understand why we have so many leaks.
The strange thing is - we have an Agilent 1200 Quaternary HPLC with the same mobile phase - it never leaks.
Could they have a raw materials issue with their stainless steel?
Regards,
Schpen
Does anyone have experience with Agilent pumps rusting? I saw one post here in 2008. We've had two pumps now with issues.
I am seeing a reddish discoloration on our pump head, outlet check valve threads. We also are getting a lot of leaks from the pump and AS capillaries. I do not recall ever having to change a stainless capillary as often as I have been this past year (at least 30 times - same and different caps).
We are running fairly conc. TFA (per Agilent Method Dev. folks) - 0.1% TFA/H2O (pump A)and 0.4%TFA/AcN (pump B).
Agilent has replaced one pump with a demo unit and after 3 months this one is experiencing leaks (3 in the past month). As we are under service contract.
Agilent has been extremely slow to respond - and do not understand why we have so many leaks.
The strange thing is - we have an Agilent 1200 Quaternary HPLC with the same mobile phase - it never leaks.
Could they have a raw materials issue with their stainless steel?