Jagged IC Peaks
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
4 posts
Page 1 of 1
I'm using a Dionex Ion Chromatography system and have just started having jagged sulfate peaks. I'm using their Seven anion standard and an AS4A column with carbonate eluent. I've been using this system for years and years and have never seen this. I've swapped out the column and that made no difference. Nor did new eluent and standard prep. I'm really curious as to why it's only on the 7th peak? Retention time and % Recovery are consistent. Any suggestions?
-
- Posts: 16
- Joined: Tue Feb 21, 2006 3:23 am
Hi JHoward333, please e-mail me your Chromeleon backup file (CMB) containing the affected sequences and we'll take a look at it (cannot make a conclusion from your description).
In the meantime,
What does your blank injection look like (running same PGM)?
Have you tried a different 7 anion standard from a different lot #?
Any other changes in the inline consumables (suppressor, etc.)?
In the meantime,
What does your blank injection look like (running same PGM)?
Have you tried a different 7 anion standard from a different lot #?
Any other changes in the inline consumables (suppressor, etc.)?
John Lim, Ph.D.
john.lim@thermofisher.com +1-408-203-2980
Manager, Global Technical Support
Unity Lab Services
A Part of Thermo Fisher Scientific
john.lim@thermofisher.com +1-408-203-2980
Manager, Global Technical Support
Unity Lab Services
A Part of Thermo Fisher Scientific
-
- Posts: 16
- Joined: Tue Feb 21, 2006 3:23 am
Update for our Chromforum colleagues: after reviewing the chromatograms, the background anion conductivity trace showed even pulsation pattern (~0.2 min baseline width/pulse @ 2.0 mL/min, ~0.1 uS intensity/pulse).

Moreover, the chromatogram of the 7 anion standards injection showed for sulfate and phosphate the baseline pulsation that rode on the upslope/downslope of the analyte peaks and gave a bumpy pattern at the peak apexes. Good RT and injection area reproducibility reported.

Conclusion: most likely a sticky check valve for this DX-500 pump. Last PM was approximately a year ago. The blank conductivity trace looks like a classical pump pulsation effect from a malfunctioning/worn check valve. There are of course other possibilities
...JHoward333 will perform a PM on the pump and report back to us. Good luck!

Moreover, the chromatogram of the 7 anion standards injection showed for sulfate and phosphate the baseline pulsation that rode on the upslope/downslope of the analyte peaks and gave a bumpy pattern at the peak apexes. Good RT and injection area reproducibility reported.

Conclusion: most likely a sticky check valve for this DX-500 pump. Last PM was approximately a year ago. The blank conductivity trace looks like a classical pump pulsation effect from a malfunctioning/worn check valve. There are of course other possibilities


John Lim, Ph.D.
john.lim@thermofisher.com +1-408-203-2980
Manager, Global Technical Support
Unity Lab Services
A Part of Thermo Fisher Scientific
john.lim@thermofisher.com +1-408-203-2980
Manager, Global Technical Support
Unity Lab Services
A Part of Thermo Fisher Scientific
-
- Posts: 3
- Joined: Thu Apr 21, 2011 8:00 pm
Thanks, John! Our conversations have been most helpful. I will perform the maintenance this week and will post how things go. Thank you for the suggestions.
4 posts
Page 1 of 1
Who is online
In total there are 380 users online :: 2 registered, 0 hidden and 378 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot], Semrush [Bot] and 378 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot], Semrush [Bot] and 378 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science