The base signal is higher 30 pA, sometimes higher 50 pA. The base line goes up and down routinely each 15 minutes with a 6-pA-drift for each.
I don't kown how to correct it.
Help me.
Advertisement
Discussions about GC and other "gas phase" separation techniques.
Hi , sorry from "poiging inglish" , naturally Russian langvige.Do you have a carbon trap between the compressor and the back of the GC? The oilless compressor compresses whatever is in the air - including hydrocarbon contaminants. Do you have a pressure regulator between the compressor and the instrument? You need to know that the pressure coming to the back of the GC is at least 20 psi above the pressure set for the air inside the instrument - and is fairly constant. Be sure you have adequate pressure drop across regulators and a constant pressure soruce for fuel gas as well.
You have answered your own question - there is something wrong with the air system, and you need to fix it.I hace a zero air generator for the compressor so the air isn't hydrocarbon and water residue. I also have a pressure generator between the compressor and the instrument.
The signal goes down as soon as the oil-less compressor works on to make up for the lack pressure.
Hi Don HiltonAlso, be sure of the tubing you have used for putting the system together. I have seen people use tubing that has not been cleaned for GC application. Such tubing may have a high background of oils in it, I presume from manufacturing.