HP 5890 + Chemstation B 04.01
Posted: Thu Apr 14, 2011 7:29 am
by intereco
I have HP 5890 series II which successfully works with Chemstation A 10.01. Now I tried to upgrade Chemstaion to B 04.01 but with this version can't esteblish stable connection with chromatograph. In Logbook Instr1.log every 3 second I see:
Code: Select all
2758 1da4 4da6a25f 300
5890 I/O error with GC: Switch to offline. 10:29:35 04/14/11
2759 1da4 4da6a25f 300
5890 Recovered from I/O Error: Back Online. 10:29:35 04/14/11
2758 1da4 4da6a262 300
5890 I/O error with GC: Switch to offline. 10:29:38 04/14/11
2759 1da4 4da6a263 300
5890 Recovered from I/O Error: Back Online. 10:29:39 04/14/11
and on hp 5890 display I see that parameters passes again and again... without stop.
new chemstation can't correctly works with hp 5890?
thanks,
-vadym
Re: HP 5890 + Chemstation B 04.01
Posted: Thu Apr 14, 2011 9:43 am
by DSP007
I have HP 5890 series II which successfully works with Chemstation A 10.01. Now I tried to upgrade Chemstaion to B 04.01 but with this version can't esteblish stable connection with chromatograph. In Logbook Instr1.log every 3 second I see:
Code: Select all
2758 1da4 4da6a25f 300
5890 I/O error with GC: Switch to offline. 10:29:35 04/14/11
2759 1da4 4da6a25f 300
5890 Recovered from I/O Error: Back Online. 10:29:35 04/14/11
2758 1da4 4da6a262 300
5890 I/O error with GC: Switch to offline. 10:29:38 04/14/11
2759 1da4 4da6a263 300
5890 Recovered from I/O Error: Back Online. 10:29:39 04/14/11
and on hp 5890 display I see that parameters passes again and again... without stop.
new chemstation can't correctly works with hp 5890?
thanks,
-vadym
Zdravstvi Vadim
1)
http://www.anchem.ru/forum/read.asp?id=12288
2) Interlab in Moskow. Best to turn to the distinguished Boris P Lapin
Lapin@interlab.ru , He has friends in Kiev. Maybe will throw the program or manual.
То же но на русском.
Здравствуй.
1) На Анхеме терли эту тему. Один из вариантов- использовать Унихром.
2) Обратись в Московский интерлаб, interlab@ sevice.ru. Лучше всего к уважаемому Борис Петровичу Лапину,
Lapin@interlab.ru. У него есть знакомые в Киеве, занимающиеся дистрибуцией и сервисом, подскажет телефончик. А может и поодкинет мануал или программный плагин из своих запасов.
Re: HP 5890 + Chemstation B 04.01
Posted: Thu Apr 14, 2011 9:57 pm
by GasMan
If you look into the 'Install' folder of your ChemStation CD and read the 'readme.txt' file you will see that the 5890 is not supported in rev B.04.01.
Gasman
Re: HP 5890 + Chemstation B 04.01
Posted: Fri Apr 15, 2011 8:02 am
by intereco
yes, I see that but there indicated:
Code: Select all
12. 5890A is no longer supported
---------------------------------
The 5890A is no longer supported by the ChemStation.
and I have hp 5890 Series II...
however, I find possible reason of problem - Min. Firmware Revision:
my HP5890 has only A.02.01 revision... as far I understand it's not simle task - upgrade firmware..?
thanks a lot,
-vadym
Re: HP 5890 + Chemstation B 04.01
Posted: Fri Apr 15, 2011 7:45 pm
by Consumer Products Guy
Updating firmware is do-able, but a little awkward. Details and firmware can be downloaded from the Agilent.com
Re: HP 5890 + Chemstation B 04.01
Posted: Sat Apr 16, 2011 10:12 am
by intereco
unfortunately, I can't find firmware for hp5890 on agilent site and wherever in internet, only for latest chromatographs... seems agilent believes that hp5890 already hopelessly obsolete...
did you meet it elsewhere?
thanks,
-vadym
Re: HP 5890 + Chemstation B 04.01
Posted: Sat Apr 16, 2011 1:42 pm
by Consumer Products Guy
I'll be back at work Tuesday. If you post your E-mail address (e.g. abd -at-134.com) I can E-mail to you all the files I have already downloaded and saved. Some of these files I think are also on the ChemStation CD disc.
Re: HP 5890 + Chemstation B 04.01
Posted: Tue Apr 19, 2011 3:10 pm
by intereco
the reason of my problem was regional setting issue of my PC (with Ukrainian options). It should be set up to English (US), or, at least, following options:
• Decimal symbol = . (point)
• Digit grouping symbol = , (comma)
• List separator = , (comma)
seems as chemstation at startup unabled set some parameters it's called error which reconnect chromatograph again and again.
After choosing correct setting on PC HP5890 and Chemstation works like a charm!
Thanks a lot,
-vadym
Re: HP 5890 + Chemstation B 04.01
Posted: Wed May 18, 2011 5:04 pm
by bgranot
I have HP 5890 series II which successfully works with Chemstation A 10.01. Now I tried to upgrade Chemstaion to B 04.01 but with this version can't esteblish stable connection with chromatograph. In Logbook Instr1.log every 3 second I see:
Code: Select all
2758 1da4 4da6a25f 300
5890 I/O error with GC: Switch to offline. 10:29:35 04/14/11
2759 1da4 4da6a25f 300
5890 Recovered from I/O Error: Back Online. 10:29:35 04/14/11
2758 1da4 4da6a262 300
5890 I/O error with GC: Switch to offline. 10:29:38 04/14/11
2759 1da4 4da6a263 300
5890 Recovered from I/O Error: Back Online. 10:29:39 04/14/11
and on hp 5890 display I see that parameters passes again and again... without stop.
new chemstation can't correctly works with hp 5890?
thanks,
-vadym
Go to
www.5890.com
you may get some help, but I am not sure