Hello i need some help how i can calculed loq and lod in lc ms ms

Thanks.
LOD is the limit of detection. The lower limit where you can say that you have a signal that stands out from the baseline noise to such a degree that you can say with a certain confidence that a peak is present.
Depending who you talk to this is usually 2 or 3 times the Signal to Noise ratio (S/N).

LOQ, Limit of Quantification, is higher. 10x S/N will usually give a high degree of method reliabilty at the low end though I have seen methods run with acceptable LOQ of 5x S/N.

Alp
This is a very good article on method validation.
http://www.ncbi.nlm.nih.gov/pubmed/16781833

Forensic Science International
Volume 165, Issues 2-3, 17 January 2007, Pages 216-224

Regards,
/Rune
see also ICH's Q2(R1):

http://www.ema.europa.eu/docs/en_GB/doc ... 002662.pdf

Although this document is greatly undermined by the mischievious inclusion of the idea of validating LOD by "visual evaluation" (i.e. "I think I can see something so it must be OK"?)
4 posts Page 1 of 1

Who is online

In total there are 55 users online :: 2 registered, 0 hidden and 53 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Amazon [Bot], Semrush [Bot] and 53 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry