What could be the reason for this HPLC analysis?
Posted: Wed Apr 13, 2011 3:37 am
Last Thursday, I prepared a standard mixture of 5 compounds of with known concentration (50 ug/mL) in MeOH and optimized a method development.
The compounds are
(1)
(2)
(3)
(4)
(5)
The method is
SP: C18
MP: H20 and ACN
Flow rate: 1ml/min; pressure limit=250
Gradient program is
At 0 min, 76% H20 & 24% ACN
At 8 min, 70% & 30% ACN
At 13 min, 64% H2O & 36% ACN
Total time=13 min; post time=3 min
The order of elution and RT are:
(2)
(1)
(3)
(5)
(4)
Today, using the same EXACT method I analyzed SPE extraction of the same 5 compounds dissolved in H2O (5 ug/mL each). I used 50ml of the aqueous std sample and used 5 ml MeOH in extraction. So, it is 50 ug/mL each.
When I analyzed it with the same EXACT method, the retention times for all 5 compounds were much shorter with the last peak coming out at 8.129 min. Here are the RT:
(2)
(1)
(3)
(5)
(4)
Note that H2O used in MP is supposed to have 1% MeOH in it to prevent mold forming. Noticing that the bottle containing H2Ois full, the prof said that whoever refilled it may have put more MeOH in it.
Right away, I tried to adjust the second part of MP in gradient program like show below (giving total run time to be 9 mins; post time 3 min) but the last peak didn't come out until 1 1/2 min after post time. It came out at 13.5 min. RT for others were changed too though the first part of the gradient program was kept the same.
Gradient program is:
At 9 min, 69% H2O & 31% ACN
Total time=9 min; post time=3 min
RT are:
(2)
(1)
(3)
(5)
(4) (didn't come out during the analysis but I saw a peak coming out 1 1/2 min after post time. I took a screen shot.
So I tried adjusting the gradient program again like this:
Gradient program is
At 12 min, 65% H2O & 30% ACN
Total time=13 min; post time=4 min
(My logic of keeping ACN to be 30% is that in the previous run, the last peak came out 1 1/2 min after post time of 3 mins, meaning the MP had gone back to original strength? I know that I should have given post time to be 5 mins
RT are:
(2)
(1)
(3)
(5)
(4) (didn't come out during the analysis and I don't know whether it did during post time or after ward because I wasn't near the computer.
Any suggestion what I should do to shorten the analysis time?
Note: The first part of the gradient program, i.e. form 0 min to 8mins can not be changed to ensure that the critical pair, namely A and B , come out with Rs>1.5.
Tomorrow, I would like to improve this method to shorten analysis time before I run my unknown. I would also re-analyze the first standard prepared using MeOH (without SPE extraction). May be I should re-analyze that first tomorrow using the same exact optimized method that gave good separation between A and B tomorrow before analyzing the SPE extracted standard mixture and SPE extracted unknown to determine whether the RTs have changed for the 5 compounds. IF the RTs have changed, that means so changes that I cannot control has happened between last Thursday and now.
Any idea how I can shorten the analysis time given the current info.
The compounds are
(1)
(2)
(3)
(4)
(5)
The method is
SP: C18
MP: H20 and ACN
Gradient program is
(2)
(1)
(3)
(5)
(4)
(2)
(1)
(3)
(5)
(4)
RT are:
(2)
(1)
(3)
(5)
Gradient program is
RT are:
(2)
(1)
(3)
(5)