CO/CO2 analysis with methaniser /FID
Posted: Tue Apr 12, 2011 11:20 pm
In an ethylene plant that requires very low levels of CO/CO2 we have both on-line and lab GC with methaniser /FID detection.
We consistanty see 200 ppb CO2 on lab test but only 20 ppb on on line analyser, when both systems are calibrated to the same 2ppm standard. obviously something strange is going on at the lower levels but I dont know if the on line analyser is giving false low results or the lab giving high results. Ideally a certified standard in the ppb range would help me resolve it but no one seems to make such low standards. To take a sample to the lab, we use a stainless steel cyclinder with quick connect valves. I dont believe we are getting contamination here. Someone suggested carrier gas purity can effect the result but I dont understand how. if there was CO2 in the lab carrier gas then wouldnt this just shift the baseline rather than contribute to the peak from the sample? Any ideas appreciated.
Terry
We consistanty see 200 ppb CO2 on lab test but only 20 ppb on on line analyser, when both systems are calibrated to the same 2ppm standard. obviously something strange is going on at the lower levels but I dont know if the on line analyser is giving false low results or the lab giving high results. Ideally a certified standard in the ppb range would help me resolve it but no one seems to make such low standards. To take a sample to the lab, we use a stainless steel cyclinder with quick connect valves. I dont believe we are getting contamination here. Someone suggested carrier gas purity can effect the result but I dont understand how. if there was CO2 in the lab carrier gas then wouldnt this just shift the baseline rather than contribute to the peak from the sample? Any ideas appreciated.
Terry