Posted: Sat Apr 09, 2011 2:06 pm
by Edu
Hi everyone,

I have been working with an Shimazdu ion chromatograph with suppressor cartridge and a mobile phase of NaHCO3/NaCO3. I use Shimazdu Column with pre-column.
Recently the pressure of my system has increased too much. I have identified that one of the cartridges suppression system appears to be blocked. I tried to run water and mobile phase at a low flow to see if I can reduce the pressure, but I have not had a good result. Before this happened, the system was working properly and a moderate pressure (70 kgf/cm2) and no changes in mobile phasehas been done at any time as to produce a precipitation in the suppressor cartridge. I wanted to see if anyone has any idea how to recover the cartridge? and what could be the cause of such obstruction?
Thanks in advance

Re: Problems with suppressor cartridge

Posted: Sun Apr 10, 2011 8:48 pm
by DSP007
Hi
You can crushed sorbent in a column or cartridge at high pressure?
Try to demotage the cartridge and set the column opposite.
If the pressure in this fall - then 100% crushed sorbent and scored sorbent filter. Column and cartridge "kaput" :cry: .