I found this application for catecholamines:
Separation of Seven Catecholamines by HPLC
A mixture of seven catecholamines was separated using a GROM-SIL 120 Octyl-5 CP column and detected by electrochemical detection.
Catecholamines II, Grom Application Number 01027
A mixture of seven catecholamines was separated using a GROM-SIL 120 ODS-4 HE column (150 x 4mm, Part No. GSOD40512S2504) by isocratic elution with a mixture of 6.9g sodium phosphate, 37mg EDTA, 150mg sodium octane sulfonate, 60mL acetonitrile, 5mL of THF and the rest made up to 1000mL with water in around 20 minutes and detected by electrochemical detection. The catecholamines separated were noradrenaline, 3,4-dihydroxyphenylacetic acid, 3,4-dihydroxyphenylalanine, dopamine, epinine, homovanillic acid and 3-O-methyldopa.
Beside the RP column what is the difference to your methode? Please let us know, and do after the first injection a blank run, just inject mobile phase. Than we will see. On your second chromatogram it seams to be a blank run, there is no separation, no compounds.