on injection of a pure sample of the above, i get several peaks. Could anyone tell me how to do away with this problem?
Advertisement
Discussions about GC and other "gas phase" separation techniques.
So what could the usefull information be?The other peaks are probably from the 60% of the sample that is not glyoxal, or if you are working with the commercial 40% aqueous solution, possibly hydrates.
Before you worry about "doing away with this probelm" you need to think carefully about whether the extra peaks are telling you something useful about the sample.
Peter
please provide information
edit: instrument manufacturer/model, column type, what kind of inlet, detector, oven parameters, flow rate, etc.
"So what could the usefull information be?" Presumably something connected with why you are running the analysis in the first place. What is the purpose of the analysis ?, what are your trying to find out that you do not know already ? What are you trying to measure ?So what could the usefull information be?The other peaks are probably from the 60% of the sample that is not glyoxal, or if you are working with the commercial 40% aqueous solution, possibly hydrates.
Before you worry about "doing away with this probelm" you need to think carefully about whether the extra peaks are telling you something useful about the sample.
Peter
I know that Glyoxal is very unstable.
It polymerizes easily.
If this is the case, then how can I run a sample of the Glyoxal without it getting to change to another compound?
"So what could the usefull information be?" Presumably something connected with why you are running the analysis in the first place. What is the purpose of the analysis ?, what are your trying to find out that you do not know already ? What are you trying to measure ?So what could the usefull information be?The other peaks are probably from the 60% of the sample that is not glyoxal, or if you are working with the commercial 40% aqueous solution, possibly hydrates.
Before you worry about "doing away with this probelm" you need to think carefully about whether the extra peaks are telling you something useful about the sample.
Peter
I know that Glyoxal is very unstable.
It polymerizes easily.
If this is the case, then how can I run a sample of the Glyoxal without it getting to change to another compound?
Peter
Yes, I acknowledge that the title was wrong. It would have been running pure Glyoxal on a GC.You can't.
Your title is a misnomer "GC analysis of pure glyoxal". This doesn't happen. By its very nature it has impurities. All you can do is see what oligomers exist and approximately how much.
You can purchase the di and tri oligomers as markers: (CAS# 4845-50-5) (CAS# 4405-13-4).
best wishes,
Rod
Thanks so much for your response!You may not be able to calibrate your results to hard figures, but you should be able to trend the results nicely, and if you are looking for the optimum reactions conditions that should suffice.
You can estimate the amounts by assuming your peak of interest will have a similar FID response to another analyte of your choosing, an external standard. Acetaldehye comes to mind as a possibility. You may choose another more suited to your needs.
I hope this is helpful
Rod
Thanks once more for your submission on this issue.This analyte and most aldehydes are very reactive. If water is in the mix, even more so.
Just injecting the aldehyde in a heated injector may induce reactions which may not be quanitative but may be reproducible.
I would investigate IC and wet chemistry methods to assist you in your research, as well as the possibility of headspace analysis.
best wishes,
Rod
Thanks once more for your submission on this issue.This analyte and most aldehydes are very reactive. If water is in the mix, even more so.
Just injecting the aldehyde in a heated injector may induce reactions which may not be quanitative but may be reproducible.
I would investigate IC and wet chemistry methods to assist you in your research, as well as the possibility of headspace analysis.
best wishes,
Rod
I patiently wait for you as it appears you are ready to offer a truce on this dillema.
thanks once more.
Never heard from you again on this issue.please provide information
edit: instrument manufacturer/model, column type, what kind of inlet, detector, oven parameters, flow rate, etc.