Need help on chromatograph interpetation

Posted: Mon Apr 04, 2011 10:42 pm
by PinotNoir
Recently I have trouble on the GC-MS sensitivity, so I just cleaned the ion source in MS, and cleaned inlet part, however, after I increase the sensitivity I have new trouble in peak shape and baseline, you can see high baseline and baseline drift and also peak tailing very badly.
Thanks for your time.

Re: Need help on chromatograph interpetation

Posted: Tue Apr 05, 2011 11:30 am
by Don_Hilton
Peak shape can be the result of a problem in the inlet or gc column - the particular bad peak shapes are diagnostic for the problems encountered. Check the websites of the suppliers of parts and supplies for GC for a troubleshooting guide. Years ago I had a small booklet from Supelco that had a fairly throuough description. It may have been long enough ago that it covered only packed GC columns -- because that's all there were "back in the day." :roll:

Baseline issues, again, have many causes. They result from either material slowly leaving the column (stationary phase bleeding off, residue from previous injections, etc.), detector issues, or electronics problems.

A bit more detail is requried to help sort this one out. I would suggest starting with a troubleshooting guide. You may recognize a problem and fix it faster than you can type a more detailed description. I will look to see either your success or more details.

Re: Need help on chromatograph interpetation

Posted: Tue Apr 05, 2011 5:41 pm
by PinotNoir
update:

I tried cleaned the TDU inlet yesterday, take them apart, cleaned and put them together, change new Cis liner, rerun it , the question is: without sample, just run empty tube(blank), the baseline is not noisy or high, so I double the problem of column bleeding, however, when I tried to run the sample, problem comes again. Compare to precious run it exactly looks same, I will change a wax column instead of this HP-5 column to check it, make sure what's going on.

If I successfully fix the problem I will update more.

Thanks

Re: Need help on chromatograph interpetation

Posted: Wed Apr 06, 2011 5:50 pm
by PinotNoir
update:

Finally I solve the problem, it from the installation of column, one session is not connect very good, so after I reinstall the column, chromatograph finally looks beautiful. Thanks guys, and have a good day.

Re: Need help on chromatograph interpetation

Posted: Fri Apr 08, 2011 7:47 am
by Peter Apps
Thanks for the feedback.

Peter