I'm doing my bachelor's thesis and i am supposed to develop a GC-method for analyzing mono-, di- and triethylene glycol in water samples within 1-100ppm.
i'm using an MMI GC-FID with a HP-5 column. i have tried Stabilwax with poor results. i have also tried 2 different types of liners (2mm dimpled liner and one with glass wool). i'm currently injecting 25ul to be able to get peaks on 1ppm.
My problem is that the liner with glass wool gives a large carry-over due to flashback, and the dimpled liner doesn't give any results on lower concentrations but large deformed peaks on 100ppm.
i have tried lower injection volumes to get rid of the flashback, but then i loose a lot of sensitivity. anybody with experience with MMI who can help me?
the company i'm doing it for doesn't want any sample preparations if it can be avoided, just take the water sample and inject it as it is.
Any pointers will be much appreciated