HP 1090 leak shutdown

Posted: Wed Mar 30, 2011 1:36 pm
by faris
Hi, i have a leak shutdown in my HP 1090. every thing is running good for about 1 min and later on I have a leak shutdown. Any help :(

Re: HP 1090 leak shutdown

Posted: Wed Mar 30, 2011 6:48 pm
by Consumer Products Guy
If you mean a fluid leak, then check the sensor in the column compartment, most leaks occur there. These must be dry; if wet, soak up the fluid and let air dry, then turn on again.

Second would be to make sure the autosampler isn't leaking at its needle/seat or at its rotor valve, if so, check its leak sensor as above.