Advertisement

HP 1090 leak shutdown

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
Hi, i have a leak shutdown in my HP 1090. every thing is running good for about 1 min and later on I have a leak shutdown. Any help :(
If you mean a fluid leak, then check the sensor in the column compartment, most leaks occur there. These must be dry; if wet, soak up the fluid and let air dry, then turn on again.

Second would be to make sure the autosampler isn't leaking at its needle/seat or at its rotor valve, if so, check its leak sensor as above.
2 posts Page 1 of 1

Who is online

In total there are 9 users online :: 1 registered, 0 hidden and 8 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Semrush [Bot] and 8 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry