Detector

Posted: Sun Mar 27, 2011 10:31 am
by EmpChrom
If you have multiple detectors in a stack how do you tell Empower which detector to collect the signal from?

Re: Detector

Posted: Mon Mar 28, 2011 1:41 pm
by DR
In the instrument method editor, click on something you do need (the buttons representing the modules) then just right-click on one you do not need, and click on "delete instrument". Repeat until just the components you want remain.

Re: Detector

Posted: Mon Mar 28, 2011 7:06 pm
by JGK
If you have multiple detectors in a stack how do you tell Empower which detector to collect the signal from?
The easiest way is to have all of the possible HPLC configurations entered in the instrument list. Then when setting the project up, specify that configuration when prompted for it