Rod thanks for fast replay
As I told my TCD is connected to the photoreactor and then I have injection by autosampler but for the experiments I mentioned I used a headspace vial i which I run the reaction and then I was sampling a gas sample by syringe and I injected manually. Injection on the column was 1ul. In case of the valve port openings I have no idea what you mean and where to check it, sorry... I have just such information on the delivery sheet: 221-VIED46UW sampling valve vici, including 6port 2-pos valve, m-electric 4"so 1/8".75mm, 330C/300psi gas, N60/T (ED46UWT), Sample loop 500 ul, UW type valve 1/8" ends with nuts and ferrules, 316SS (SL500UW) - do you understand something form this? otherwise I will call Shimadzu and ask them. ID of my column is 2.1 mm (0.08 inch) it's pack Porapack Q 80/100 3m column. In my case doesn't matter the parameters I have seen just tailing of big air peak, CO2 on the tail and no water. Just when I froze out the water (immersed few time the vile in liquid nitrogen) I have seen good separation of air and CO2 and no water peak. Water peak I could observed when I replaced Porapak Q by Carboxen 1000 but then water and CO2 were coming as a double peak and could not resolve it. My support person from Shimadzu told me that I am "killing" my Porapak Q column by water and that I will not get separation on this column if I have water in my sample. I also had a meeting with expert from Supelco and he told "GC is not happy with water and you are not happy with your GC"

Therefore I was research if I can remove water before injecting on the column or somehow before loop but then it is always danger that I will remove CO2 as well. Do you think that I can use a Porapak Q for water separation? The other alternative I found are PLOT columns like PoraPlot Q or U or HayeSep D but this require reconstruction of GC.
Thanks for support
Kasia