Karen01 » Tue Apr 19, 2011 4:30 am
													
 
						
					 
					
						
I once did an analysis looking for ppm levels of hydrocarbons (pump oil actually) from C16 to C37 in a water solution (the matrix acted like a soap so the extraction was difficult).
Yes, I am concerned about  emulsions... That is why i was originally thinking of starting with a RF SPE cartridge and wash with a low MeOH water solution, elute off with hexane and run that through a drying cartridge ... But not sure if that would work and is a lot of manipulation ..and even if it did it would probably require some evaporation.
I used a 3 meter 0.53mm ID OV-17 1 micro film column for this analysis, running from 40 to 275C  at 30C/min.  FID   One 30m commercial column became 10 test columns.
I will look into that column. Almost all my GC experience up until now was for residual solvents. This is a whole different kettle of fish!
I extracted the aqueous solution with hexane and CAREFULLY with a gentle flow of nitrogen blew down the several mL of hexane to 0.25mL or less.  Internal std is recommended unless you can measure small amounts of the final volume accurately.
How did you deal with the emulsion issue in the extraction?
I did not dry the solution as the water collected nicely on the inside of the culture tube I used to hold the last few mL of solvent.
Purging with dry nitrogen through a needle in a vial with a septum (preferably a "v" vial) that was vented  with a needle could prevent that. That could be done on a manifold so 6- 8 could be done at time. But as hood space is at real premium i would like to avoid having to remove solvent that way.
I think I injected 2 microliters SLOWLY into a injection liner DIRECTLY attached to the column (NO SPLIT)
In theory the Muti-mode inlet should let me inject up to about 50 uL and as I will be able to program l control the  Inlet temperature, i should be able to selectively evaporate the hexane first there. I've never used this inlet so I'll have to see what it can really do.
Just a FYI, it can be done, and I did no drying of the samples.
Being able to avoid the drying step would help for sure! All the environmental assays I've seen do it as then then separate the aromatics from from the aliphatic hydrocarbons on a florisil column before injecting on GC.
Thanks,
- karen