SEC question - poor chromatography
Posted: Tue Mar 22, 2011 7:17 pm
Virtual Colleagues,
A co-worker is trying to develop a SEC method (at which we, the analytical group, are novices) to detect/quantitate a PEG acid molecule. The compound consist of an acid group at each end of two cross bar PEG units. i.e., an X-shaped molecule. The MW is ~ 40K.
The LC conditions reported to me are:
GFC BioSep-SEC-s3000 column, 300 mm x 7.8 mm
Mobile phase: 54 mM HOAc pH 3 (checked w/ pH meter)
Flow rate: 1 mL/min
Inj volume: 50 uL
Col temp: 40 C
Sample conc.: 200 ppm
Sample diliuent: water, or mobile phase
Detector: ultra CAD; neb temp: 35 C, Range: 100 pA, Filter: none
Issues: inconsist chromatography for same sample on consecutive injections, small negative spikes in the baseline, loss of peak response over time (co-worker states it is not a degradation issue).
Not sure if these are instrument or chemical related.
Thanks in advance,
R.J.
A co-worker is trying to develop a SEC method (at which we, the analytical group, are novices) to detect/quantitate a PEG acid molecule. The compound consist of an acid group at each end of two cross bar PEG units. i.e., an X-shaped molecule. The MW is ~ 40K.
The LC conditions reported to me are:
GFC BioSep-SEC-s3000 column, 300 mm x 7.8 mm
Mobile phase: 54 mM HOAc pH 3 (checked w/ pH meter)
Flow rate: 1 mL/min
Inj volume: 50 uL
Col temp: 40 C
Sample conc.: 200 ppm
Sample diliuent: water, or mobile phase
Detector: ultra CAD; neb temp: 35 C, Range: 100 pA, Filter: none
Issues: inconsist chromatography for same sample on consecutive injections, small negative spikes in the baseline, loss of peak response over time (co-worker states it is not a degradation issue).
Not sure if these are instrument or chemical related.
Thanks in advance,
R.J.