I am new to running the GC/MS. I am running EPA method 8260 for Volitale compounds in wastewater. I have noticed that we have a consistent step in our chromatograph around the 5-6 min mark. it is only elevated 60000 response counts, but this seems to be enough to be altering our data for our early eluting compounds. I have attached an image and can provide more if needed. Please let me know if you have a suggestion as to why this is occurring and how I can correct it. Also, I hope I posted this in the correct board. Thanks
Image
That should be water peak. You can reduce it by shorting your desorb time, if your method allows for it, but all your compounds will be affected as well. If it´s not posing a major interference I would ignore it...
ERA_Erin,

At 4-6 minutes, the peak suggests you have about a 3 minute desorb? Shortening the desorb could help. However, you should not be scanning so low that you would see water or even CO2 for that matter. So the first question is, what is your scan range? Can you post a spectrum of that "peak". Also, you seem to have a fairly high noise level relative to the internal standard peaks. Your scan range is a factor here as well but so is your split ratio. Can you post that as well?

Best regards,

AICMM
3 posts Page 1 of 1

Who is online

In total there are 63 users online :: 0 registered, 0 hidden and 63 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: No registered users and 63 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry