You originally mentioned "according to the literature". That implies that you are following a published, validated method. If it has been validated and you are following it *exactly* then you should not have any problems. By "exactly", I mean just that, including using exactly the same kind of column and with your instrument set up exactly the same way.
If you are changing things, you can anticipate a number of *possible" problems:
- 10 mM is a *very* low buffer concentration for a 200 microliter injection of pH 1.8 sample. If I were setting up a method, I would use 100 mM. That said, if it's a validated method and it works, then it works!
- Given the big disparity between what you are injecting and your mobile phase, I don't think *any* amount of mixing volume will help, so don't worry about that aspect. Again, if it works, it works.
- At low pH, the bonded phase typically comes off; that reduces the amount of C18 and increases the amount of silanol activity. The most common symptom is peak shape problems. Retention usually decreases (bot sometimes increases -- depends on whether the analytes interact more strongly with the C18 or the silanols!). Generally speaking, column lifetime problems start to show up with mobile phase pH values below about 2.5, but there are columns that are designed for lower pH operation (that's why it's important to follow a successfully validated method *exactly*).
You will get the longest lifetime if you flush the column with buffer-free mobile phase for overnight (or longer). The price you pay is that you will have to re-equilibrate the column with the buffered mobile phase before you can run again.
My guess is that if the method works, you *will* have a short column lifetime, but remember that the column *is* a consumable item. Even if you only get 3-4 weeks and/or a couple of hundred injections from the column, the bulk of your analysis cost will still be operator time.
"Columns are cheap; your time is valuable."