Hi

I would like to know what all precautions (or changes to system) should I take of using regular Agilent 1200 series HPLC with mobile phase pH of 9 ??

Thanks
Non

Best Regards
Learn Innovate and Share

Dancho Dikov
will there be no effect on the tubings and pump??
Per a document on Agilent's website:

"In general, 1200 modules operate over a pH range 1 to 12.5.

pH > 9.5
• Replace standard (Vespel) rotor seals in all rotary valves with either
Tefzel or PEEK seals.
• Replace the standard glass solvent inlet filters with stainless steel inlet
filters.
• Be aware that quartz flow cell windows are slowly etched. Don’t let
high pH solvents stand in the flow cell for long periods."

http://www.chem.agilent.com/Library/sli ... 8_Tour.pdf

So, you're at pH 9, which is right on the border of their precautions, so it might be a good idea to swap the recommended parts just to be safe.
Time flies like an arrow. Fruit flies like a banana.
4 posts Page 1 of 1

Who is online

In total there are 24 users online :: 0 registered, 0 hidden and 24 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 24 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry