HPLC Sunscreen method Validation
Posted: Tue Mar 15, 2011 3:55 pm
Hey everyone,
I've been lurking around here a lot lately and I decided I should finally make an account and post my question. I'm sort of new to this so please forgive me if I ask stupid questions.
First of all, let me introduce my situation. I'm a new hire at the in-house analytical lab of a cosmetic company that manufactures sunscreens. I was asked to develop and validate a method to assay Benzophenone-3, Octocrylene, Octyl methoxycinnamate, Butyl Methoxydibenzoylmethane (avobenzone or Parsol1789), 2-ethylhexyl salicylate, and homosalate in one injection. I am not an Analytical Chemist neither have I had any formal training in HPLC but I have educated myself by reading all relevant literature. The company already had a method in place but that method did not meet the USP and FDA requirements for capacity factors and for the resolution between Octyl methoxycinnamate and Butyl Methoxydibenzoylmethane. I tried several other methods that were published online that successfully fulfilled what I needed but for some reason I still had the same problem of not being able to resolve Octyl methoxycinnamate and Butyl Methoxydibenzoylmethane. I tried 3 different columns and various mobile phases and gradient tables to no avail. So I came here to seek experts help.
I have a Shimadzu VP system with an SIL-10AD autoinjector, two LC-10AT pumps, a degasser, SPD-10A UV-Vis detector that can detect up to two wavelength simultaneously, and a CTO-10A oven. The maintenance that I have performed so far is change the UV lamp and the needle seal.
The old method we used had an 85:15 (Methanol:0.1% Trifluoroacetic Acid) flow rate of 1.0 mL/min at 325nm on a Phenomenex Luna C18(2) 150x4.60mm 5 micron. At 26 degrees celsius with 5 microliters injection. The resolution between Octyl methoxycinnamate and Butyl Methoxydibenzoylmethane was always below 1.8
I have also tried a method using 80:20 (Ethanol: 0.8%Acetic Acid) flow rate of 0.7 mL/min at 313 and 354 nm on an Acclaim 100x4.60 mm 3 micron. At 26 degrees celsius with 5 microliters injection.The resolution between Octyl methoxycinnamate and Butyl Methoxydibenzoylmethane was always below 1.6.
The third method was me just messing around with a 4.6x50mm 2.7 micron Agilent column also to no avail.
Whenever I was able to increase resolution a bit, the peaks become too broad and distorted so I'm thinking my HPLC system has too much dead volume, or I need a smaller size detector cell. Please help in any way you can.